Dadof2boys
Posts: 11
Joined: Sat Oct 10, 2015 7:46 am

Manually Calibrate Salinity

Post by Dadof2boys »

Here's my situation... I have calibrated my salinity probe with 35 ppt refractive index standard solution and when in the solution the probe is reading 35.0. When I move the probe to my RODI water it reads 6.0 and I know that is really at 0. When I move the probe to my sump the probe then reads anywhere from 34-35.6. I know that my water is really at a 1.025-1.026.

Without just saying take 10 points off the probes value is there a way to force the adjustment?
User avatar
lnevo
Posts: 5430
Joined: Fri Jul 20, 2012 9:42 am

Re: Manually Calibrate Salinity

Post by lnevo »

6 is the lowest the probe reads. It's normal. It sounds like you're probe is reading perfectly for 1.025-1.026.
Dadof2boys
Posts: 11
Joined: Sat Oct 10, 2015 7:46 am

Re: Manually Calibrate Salinity

Post by Dadof2boys »

So how do you know what your salinity really is then? Do you just mentally subtract 10 points?
rimai
Posts: 12881
Joined: Fri Mar 18, 2011 6:47 pm

Re: Manually Calibrate Salinity

Post by rimai »

Do you keep your tank at 25ppt?
Salinity measurements are in ppt and not specific gravity.
Roberto.
User avatar
lnevo
Posts: 5430
Joined: Fri Jul 20, 2012 9:42 am

Re: Manually Calibrate Salinity

Post by lnevo »

35ppt is 1.026