Ph cal
Re: Ph cal
ok, tried to recalibrate, before seeing this, after I loaded a new memory file, as I wanted to make some changes, readings came in as follows, 7.0 - 537 10 - 825, then hit ok, and unpluged the controller, and plugged back in, same thing.
loaded your above code,
same version
550
840
7.0 - 529
10 - 819
tank - 613
loaded your above code,
same version
550
840
7.0 - 529
10 - 819
tank - 613
Re: Ph cal
That's strange.... It seems that the calibration is not saving the settings.
Let's try something else:
Add these 2 lines to your PDE file, just bellow the ReefAngel.Init()
Check the pH of tank water now. Does it look correct?
Let's try something else:
Add these 2 lines to your PDE file, just bellow the ReefAngel.Init()
Code: Select all
ReefAngel.PHMin=529
ReefAngel.PHMax=819
Roberto.
Re: Ph cal
well it was still off but after messing around with the numbers a bit I got it to read the tank and the 10.0 solution correctly , but 7.0 reads at 7.3
ended up with 502 and 817
is it a concern that its not saving the calibration in mem?
Tom
ended up with 502 and 817
is it a concern that its not saving the calibration in mem?
Tom
Re: Ph cal
Maybe you are clicking Cancel, instead of Ok?
I'm not sure why it is not saving. Did you try modifying the memory settings with the RAGen Internal Memory tab? Or with the Client 2.2?
I'm not sure why it is not saving. Did you try modifying the memory settings with the RAGen Internal Memory tab? Or with the Client 2.2?
Roberto.
Re: Ph cal
Im pretty sure im hitting ok, it looks highlighted, unless cancel is default, but ive tried both. I modified the pde, but also did the mem settings in RAGen
Re: Ph cal
This is how RA reads/writes the calibration data:
When you use the calibration function, it will store the data in the memory.
It will use this data for as long as you do not write anything over. If you use the calibration function again, it overwrites the previous data. If you use the Internal Memory sketch from RAGen, it overwrites the previous data. If you push memory settings with Client 2.2, it overwrites the data.
If you add the 2 lines of code mentioned above, it will use the data in the code and disregard any settings stored in the memory, but it will not overwrite anything. If you remove the 2 lines, you go back to use whatever was stored in the memory.
To make sure that it is indeed saving settings on the memory, you can either use the calibration function again and upload the above code right after to check if it is reading the same values as you saw on the screen while calibrating or you can use the internal memory sketch and upload the above code right after, checking for the same data you entered on the RAGen Internal Memory tab.
It should show you the same data.
When you use the calibration function, it will store the data in the memory.
It will use this data for as long as you do not write anything over. If you use the calibration function again, it overwrites the previous data. If you use the Internal Memory sketch from RAGen, it overwrites the previous data. If you push memory settings with Client 2.2, it overwrites the data.
If you add the 2 lines of code mentioned above, it will use the data in the code and disregard any settings stored in the memory, but it will not overwrite anything. If you remove the 2 lines, you go back to use whatever was stored in the memory.
To make sure that it is indeed saving settings on the memory, you can either use the calibration function again and upload the above code right after to check if it is reading the same values as you saw on the screen while calibrating or you can use the internal memory sketch and upload the above code right after, checking for the same data you entered on the RAGen Internal Memory tab.
It should show you the same data.
Roberto.
Re: Ph cal
I previously commented that I'd fix the PH calibration to take affect immediately. I just looked at the code more closely and it should take affect immediately. The values get updated when you press the OK button. They actually get written to memory and copied over right away.
curt
curt
Re: Ph cal
I am also having pH calibration issues. After following the procedure here, I test with cal solution and find:
pH 7.0 solution reads 7.0 and stable after 10 minutes
pH 10.0 solution reads 9.76 after 20 minutes and still drifting up and down by 1.
And yes I did power cycle after calibration. Dev Lib Version 0.8.5.19
Readings during calibration (stabilized):
7 - 544/548
10 - 835/851
Any suggestions?
pH 7.0 solution reads 7.0 and stable after 10 minutes
pH 10.0 solution reads 9.76 after 20 minutes and still drifting up and down by 1.
And yes I did power cycle after calibration. Dev Lib Version 0.8.5.19
Readings during calibration (stabilized):
7 - 544/548
10 - 835/851
Any suggestions?
_____
Jim
Jim
Re: Ph cal
Are you saying it is drifting by 1 whole pH unit? like 9 and 11pH?
If it is just drifiting by 0.01, then you are fine. You will never get lab grade pH readings. Even the probe is not lab grade.
Seems like the PHMin is fine according to your results. Just increase/decrease the PHMax until you get the correct reading.
You can use the Client to change those values too.
The value must be somewhere in the middle of the range you observed.
Or hard code it like this in the setup() section:
If it is just drifiting by 0.01, then you are fine. You will never get lab grade pH readings. Even the probe is not lab grade.
Seems like the PHMin is fine according to your results. Just increase/decrease the PHMax until you get the correct reading.
You can use the Client to change those values too.
The value must be somewhere in the middle of the range you observed.
Or hard code it like this in the setup() section:
Code: Select all
ReefAngel.PHMin=546
ReefAngel.PHMax=835
Roberto.
Re: Ph cal
Sorry - it's drifting between 9.75 and 9.77 and ReefAngel shows:rimai wrote:Are you saying it is drifting by 1 whole pH unit? like 9 and 11pH?
If it is just drifiting by 0.01, then you are fine. You will never get lab grade pH readings. Even the probe is not lab grade.
Seems like the PHMin is fine according to your results. Just increase/decrease the PHMax until you get the correct reading.
You can use the Client to change those values too.
The value must be somewhere in the middle of the range you observed.
Or hard code it like this in the setup() section:Code: Select all
ReefAngel.PHMin=546 ReefAngel.PHMax=835
pH 7: 544
pH10: 851
So do I increase or decrease pH10 to raise the value measured?
_____
Jim
Jim
Re: Ph cal
Not sure what is going on, but the pH calibration function appears not to function correctly. I finally had to take the initial readings given by the function and then read the pH7 and pH10 calibration fluids and manually adjust from within ReefAngel Client. It took me about three adjustments with each the pH7 and pH10 calibration fluids to tune it in. My pH probe is now in sync with my pH test kit.
Needless to say that cycling the equipment that many times adds to the possibility of failure. You really need to address this issue.
Needless to say that cycling the equipment that many times adds to the possibility of failure. You really need to address this issue.
_____
Jim
Jim
Re: Ph cal
Did you rinse the probe before transfering and let it stabilize on the pH 10 for while?
Roberto.
Re: Ph cal
They shouldn't be the same.
What happens to those numbers if you dip the probe into vinegar?
What happens to those numbers if you dip the probe into vinegar?
Roberto.
Re: Ph cal
mine was weird in that my numbers were:
211 (7.0)
523 (10.0)
Right now reads 7.82
211 (7.0)
523 (10.0)
Right now reads 7.82
Re: Ph cal
For a Standard test that is cheep and repeatable I use the 20 Mule Team Borax test.
Borax or sodium metaborate solution will always be at a pH of 9.18 + 0.02, even if you do a poor job of making this solution.
Use a pint of water (473ml) and add a half teaspoon (1/2 level tsp.) of borax to make a resultant pH standard solution with about 2 gm of sodium metaborate per 473 ml or 4.3 grams of sodium metborate/litre.
Although using this method to make the standard will only result in an accuracy of about plus or minus 10% of the lab standard sodium metaborate pH calibration solution, for me it's a good and cheap test to make sure the probe is on the mark. Compared to commercial solutions you can afford to use this test every week when you clean your probe.
Borax or sodium metaborate solution will always be at a pH of 9.18 + 0.02, even if you do a poor job of making this solution.
Use a pint of water (473ml) and add a half teaspoon (1/2 level tsp.) of borax to make a resultant pH standard solution with about 2 gm of sodium metaborate per 473 ml or 4.3 grams of sodium metborate/litre.
Although using this method to make the standard will only result in an accuracy of about plus or minus 10% of the lab standard sodium metaborate pH calibration solution, for me it's a good and cheap test to make sure the probe is on the mark. Compared to commercial solutions you can afford to use this test every week when you clean your probe.
Re: Ph cal
Add these 2 lines to your PDE file, just bellow the ReefAngel.Init()
Code: Select all
ReefAngel.PHMin=520
ReefAngel.PHMax=801
Roberto.
Re: Ph cal
sorry. forgot to add ;
use this one instead:
use this one instead:
Code: Select all
ReefAngel.PHMin=520;
ReefAngel.PHMax=801;
Roberto.
Re: Ph cal
Is there a way to change the values I calibrate to? I only have available pH 4 and pH 9 calibration solutions from all of the LFS in the area. I could order the pH 7 and pH 10 calibration solutions, but then I'd have to order the big bottle.
Re: Ph cal
It's simple interpolation math.
You can get the numbers from pH4 and pH9 and calculate both 7 and 10 values.
You can get the numbers from pH4 and pH9 and calculate both 7 and 10 values.
Roberto.
Re: Ph cal
awesome, how would I accomplish this?
Just a thought from a noob and some reading
pH 4 reading would be something like 398
pH 9 reading would be something like 903
I would just change the numbers to make them work for the pH 7 and pH 10, something like this?
adjust pH 4 reading to 698
adjust pH 9 reading to 1003
Just a thought from a noob and some reading
pH 4 reading would be something like 398
pH 9 reading would be something like 903
I would just change the numbers to make them work for the pH 7 and pH 10, something like this?
adjust pH 4 reading to 698
adjust pH 9 reading to 1003
Re: Ph cal
Correct. You just need to calculate the simple interpolation math.
Not too hard, but if you get problems, let us know.
Not too hard, but if you get problems, let us know.
Roberto.
Re: Ph cal
When I finish the pH calibration and leave the probe in the 10.0 solution the controller shows the pH @ 9.87 (which is close enough). But when when I remove the probe from the 10.0 solution and put it in my sump the reading on the controller jumped to 20.21. I put the probe back in the 10.0 solution and the display now reads 19.43. There is no way the pH in my tank is that high. I tested with a Salifert pH test and it shows 8.3
Rebooting the controller doesn't make any difference. I've repeated the calibration 4 times now with the same results.
I've only had the controller setup for just over a week now and my probe is new and never been exposed to air.
What can I do?
Thanks
Rebooting the controller doesn't make any difference. I've repeated the calibration 4 times now with the same results.
I've only had the controller setup for just over a week now and my probe is new and never been exposed to air.
What can I do?
Thanks
Re: Ph cal
Yesrimai wrote:Did you do pH7 first?
Re: Ph cal
7.00 = 243
10.00 = 360
10.00 = 360