Here are the numbers..
PH4 217
PH7 512
Hopefully you can show me how to calculate the PH10 number from this.
Remco
PH calibration
Re: PH calibration
http://en.wikipedia.org/wiki/Linear_interpolation
Let's say that x axis is pH and y axis is the raw numbers.
We know the following:
x0=4
y0=217
x1=7
y1=512
So, the formula would be:
y=217+((x-4)*((512-217)/(7-4)))
We want to interpolate and find the y value for when x=10
y=217+((10-4)*((512-217)/(7-4)))
y=807
So, in your case, you can use PHMin=512 and PHMax=807
Just place these lines just below the ReefAngel.Init() function:
Let me know if it got too confusing.
x0=4
y0=217
x1=7
y1=512
y=807
ReefAngel.PHMin=512;
ReefAngel.PHMax=807;
Re: PH calibration
An easier way for the future, for those that don't want to bother with the math part of it.
Just take the difference between pH7 and pH4, add it to pH7 and you've got your pH10 number.
512-217=295
512+295=807
807 is your number to be used for PHMax.
PHMin is always the same as pH7 raw number reading.
512-217=295
512+295=807
Re: PH calibration
Thnx for all the info !
I also have tested changing the PH.Params map. That worked also without a problem.
So now I have 2 way's of using PH4 calibration fluid.
Re: PH calibration
Just remember that the changes in the map function inside the library will get overwriten by any update that Curt releases on the libraries and you will need to change it again.