water level sensor pipe diameter

Expansion modules and attachments
Post Reply
binder
Posts: 2871
Joined: Fri Mar 18, 2011 6:20 pm
Location: Illinois
Contact:

water level sensor pipe diameter

Post by binder »

i’m gonna work on my own dosing containers and i plan on using the water level sensors. i know the 1/2” pipe is recommended but i wanted to use something different. i was hoping to use like a 1/4” or 3/16” tube with the air hose connected to it, so i have a small footprint in my container. will this work or not? i haven’t had time to test on my setup so i figured i would ask first and i didn’t find anything concrete on here.


Sent from my iPad using Tapatalk
rimai
Posts: 12881
Joined: Fri Mar 18, 2011 6:47 pm

Re: water level sensor pipe diameter

Post by rimai »

It doesn't work.
There is not enough pressure in a such small tubing.
You would need a custom module with more sensitive sensors.
Roberto.
binder
Posts: 2871
Joined: Fri Mar 18, 2011 6:20 pm
Location: Illinois
Contact:

Re: water level sensor pipe diameter

Post by binder »

ok. good to know. so the minimum pipe diameter that will currently work is the 1/2". thanks!

doing something like this with the more sensitive sensor could be something that might be of interest to people...might be something to look into pricing for it. just a thought.

anyways, thanks for the info!

Sent from my XT1585 using Tapatalk
rimai
Posts: 12881
Joined: Fri Mar 18, 2011 6:47 pm

Re: water level sensor pipe diameter

Post by rimai »

It is very expensive for a more sensitive sensor :( It costs 3 times more.
I think it would be cheaper to use etape.
Roberto.
rimai
Posts: 12881
Joined: Fri Mar 18, 2011 6:47 pm

Re: water level sensor pipe diameter

Post by rimai »

etape is even more expensive :(
Roberto.
binder
Posts: 2871
Joined: Fri Mar 18, 2011 6:20 pm
Location: Illinois
Contact:

Re: water level sensor pipe diameter

Post by binder »

rimai wrote:It is very expensive for a more sensitive sensor :( It costs 3 times more.
I think it would be cheaper to use etape.
holy cow! oh well, thanks for checking! I've got an idea for what I can do and it will just require using the 1/2" pipe.

Sent from my XT1585 using Tapatalk
User avatar
lnevo
Posts: 5430
Joined: Fri Jul 20, 2012 9:42 am

Re: water level sensor pipe diameter

Post by lnevo »

would a wider diameter work better and give us better granularity for deeper measurement? or wouldn't do anything different?
rimai
Posts: 12881
Joined: Fri Mar 18, 2011 6:47 pm

Re: water level sensor pipe diameter

Post by rimai »

For deeper measurements, you actually need the opposite.
The wider the diameter, the more displacement of air and thus more pressure you are going to generate.
You need a sensor that will read more pressure.
The one we use now is 0.57psi. For deeper measurements, you may create more pressure than that and for small tubing, you don't create not nearly enough pressure.
But, if you use a lesser diameter for a deeper measurement, it may work good for you with the current sensor.
Roberto.
User avatar
lnevo
Posts: 5430
Joined: Fri Jul 20, 2012 9:42 am

Re: water level sensor pipe diameter

Post by lnevo »

Hmm so maybe a John Guest 1/4" -> 3/8" and a 3/8" rigid tube?
rimai
Posts: 12881
Joined: Fri Mar 18, 2011 6:47 pm

Re: water level sensor pipe diameter

Post by rimai »

It should work.
Roberto.
User avatar
Pny
Posts: 35
Joined: Wed Aug 08, 2012 5:06 am
Location: Sweden

Re: water level sensor pipe diameter

Post by Pny »

The pressure in the air column inside the pipe is the same regardless of the shape and dimensions of the pipe.

Skickat från min G8441 via Tapatalk
200 liter reef cube + 1000 liter FOWLR
rimai
Posts: 12881
Joined: Fri Mar 18, 2011 6:47 pm

Re: water level sensor pipe diameter

Post by rimai »

Is it?
I think the shape and dimension is a factor.
The bigger the shape, the more air displacement per linear measurement it will have as the water rises inside the tube.
Let's say 1cm of water rise in the tube of 1/4" will displace less air and create less pressure as oppose to a 2" tube.
Or am I not understanding this correctly.
Roberto.