problem ph
Re: problem ph
kh is from 4 to 128 and down to 50
7 is for kh 230 and goes up to 100
7 is for kh 230 and goes up to 100
Re: problem ph
Let me explain by placing the probe in the solution of the reef angel 4 and going in the ra ph calibrazioen also scuotenso the probe does not give me less than 50 is put in solution 4 and 7 solution
Re: problem ph
I think for on 4 04 and on 7 07 correct?
problem ph
Those numbers you are seeing in the calibration screen aren't the actual ph numbers. They are used to calibrate the control.
Re: problem ph
then how do I calibrate the probe to 4 and 7? what numbers do I look?
problem ph
Calibrate the probe using the calibrate screen by using the 4 solution and then the 7 solution and then exit out of the calibration to the main screen. Now you can verify from the main screen by putting the probe in the 4 solution and seeing what it reads on the main screen then do the same for the 7 solution.
Re: problem ph
Take a look at this topic it might be what you are looking for.
http://forum.reefangel.com/viewtopic.ph ... tion#p2210
http://forum.reefangel.com/viewtopic.ph ... tion#p2210
Re: problem ph
how do I make the count of those numbers for ph 4 and 7?
Re: problem ph
The numbers you see during calibration are the raw 8-bit ADC value that Arduino has internally.
It's a value that goes from 0 to 1023.
ph 7 is around 540 in that range and ph10 is around 830. So, for ph4, you will be looking at somewhere around 250 if my math is correct....
It's a value that goes from 0 to 1023.
ph 7 is around 540 in that range and ph10 is around 830. So, for ph4, you will be looking at somewhere around 250 if my math is correct....
Roberto.
Re: problem ph
hello I tried to calibrate the pH and after an hour in kh 4 mi marks the maximum number of controller 86 and kh 7-329
Re: problem ph
Where should I put these two lines?
ReefAngel.PH Min = 250;
ReefAngel.PHMax = 512?
ReefAngel.PH Min = 250;
ReefAngel.PHMax = 512?
Re: problem ph
You don't need those lines.
Just go through calibration and make sure you pick ph4 and ph7 as reference or it won't work.
Just go through calibration and make sure you pick ph4 and ph7 as reference or it won't work.
Roberto.
Re: problem ph
I have to cancel these lines?
Re: problem ph
when time for calibration ph 4 and 7 in total?
Re: problem ph
there is a test to be done to see if the probe is to be changed?
Re: problem ph
Calibrate and place the probe into the calibration fluids you used in the calibration.
It should show their value.
It should show their value.
Roberto.
Re: problem ph
yes, I've done 100 times when I put it in the solution pH 4 I get kind ph 4:40
R: problem ph
Explane please
Inviato dal mio GT-I9300 con Tapatalk 2
Inviato dal mio GT-I9300 con Tapatalk 2
Re: problem ph
Maybe you ought to try the 7 and 10 if 4 and 7 is not working for you.
Someone else did it with 4 and 7 and got it to work just fine.
Someone else did it with 4 and 7 and got it to work just fine.
Roberto.
Re: problem ph
I need the probe 4 and 7 because I have to connect the reactor elletrovalvola
Re: problem ph
Linearity of pH probe is pretty good within all ranges that we use:
http://en.wikipedia.org/wiki/Glass_electrode
http://en.wikipedia.org/wiki/Glass_electrode
Roberto.
Re: problem ph
Good morning and Merry Christmas to everyone today I calibrated the probe again and me with solution 4 = 4.16 and 7 = 7.02 solution over so I can not right?
I need to update some library?
I need to update some library?
Re: problem ph
Looks pretty good.
You just need to let the probes stabilize a little more before you complete and also make sure that the solutions are the same temperature as your tank water.
You will never get exact measurements. This is not a scientific device nor the probe is scientific/lab grade.
Personally, your calibration is fine as it is.
It has a .16 deviation on a 3 ph span.
In the range of pH you are going to be working on, it would mean something like 0.025 ph deviation
I don't think your reactor can even notice this minute difference.
i more bubble inside your reactor can cause more deviation than that...
Just my 2 cents, but if you really want to push to get it down to perfect calibration, please go ahead.
You just need to let the probes stabilize a little more before you complete and also make sure that the solutions are the same temperature as your tank water.
You will never get exact measurements. This is not a scientific device nor the probe is scientific/lab grade.
Personally, your calibration is fine as it is.
It has a .16 deviation on a 3 ph span.
In the range of pH you are going to be working on, it would mean something like 0.025 ph deviation
I don't think your reactor can even notice this minute difference.
i more bubble inside your reactor can cause more deviation than that...
Just my 2 cents, but if you really want to push to get it down to perfect calibration, please go ahead.
Roberto.
Re: problem ph
good day yesterday, the solenoid valve is always on, and marks on the reef angel ph 680 up and down, now I have removed the pH probe from the reactor and put into solution pH 4 and pH 6.60 tells me, I try to recalibrate and idem, which I do?
with the reef angel...
with the reef angel...