How long can the PVC tube be or better question what is the capability of the level module to measure (min/max) inches of water? Also, I will have this set-up as an attachment to a poly vessel that will be my sump that my return pump is attached to. It is here I will maintain level, but I’m worried that when my system pumps are off and water fills the poly vessel, a fair amount of pressure may be applied to the level module…. Could be as much as 30 inches of water pushing on the level module. Should I be concern? What is the max pressure for the level module?
Re: Water Level module tube length and pressure
It can measure from 0 to 3.92 kPa
Max pressure is 16kPa.
Min pressure is atmospheric pressure (0 kPa).
Technically, it can measure a little bit of vaccum, so it could be less than atmospheric pressure, but don't count on it.
I think it is there just to account for differences in altitude, which would influence the zero mark of atmospheric pressure.
Max pressure is 16kPa.
Min pressure is atmospheric pressure (0 kPa).
Technically, it can measure a little bit of vaccum, so it could be less than atmospheric pressure, but don't count on it.
I think it is there just to account for differences in altitude, which would influence the zero mark of atmospheric pressure.
Roberto.
Re: Water Level module tube length and pressure
Roberto - good stuff... So I l can measure up to 15.7 inches of water and the pressure sensor can handle up to 64 inches. My plan should work.
Thanks.
Troy