Just wondering if anyone has any ideas as to why my PH reading a seem to sometimes read really low (like 1).
Seems to happen sometimes if I rewire things or plug new stuff in.... but even if I recalibrate the probe it still ends up reading funny. Currently reading 1.63 after just being calibrated...
It's a brand new probe (thought my old probe was the issue... now I'm no so sure).
Sent from my iPhone using Tapatalk
PH probes - reading strange values
Re: PH probes - reading strange values
Stray voltage. Try running the probe in cup of tank water to see if any different
Re: PH probes - reading strange values
ok. no idea what would be causing it. What can I do about it long term?
Sent from my iPhone using Tapatalk
Sent from my iPhone using Tapatalk
Re: PH probes - reading strange values
Hi, perhaps the probe are very near from fresh water reposition? I saw some similar problem in other post.
Regards,
Regards,
soyrex wrote:Hey all!
Just wondering if anyone has any ideas as to why my PH reading a seem to sometimes read really low (like 1).
Seems to happen sometimes if I rewire things or plug new stuff in.... but even if I recalibrate the probe it still ends up reading funny. Currently reading 1.63 after just being calibrated...
It's a brand new probe (thought my old probe was the issue... now I'm no so sure).
Sent from my iPhone using Tapatalk
Re: PH probes - reading strange values
Did you try the cup? Were the readings normal there? If so you need to identify what piece of equipment is causing it. I would start with float switches followed by return and skimmer pumps. Keep going till you find the culprit.
Re: PH probes - reading strange values
tried the cup. it's steady at 7.32 - didn't have time to re-calibrate it...but it's steady and holding.. so i guess it's a voltage issue.
I took the floats switches out.. no change. I will try and isolate cause ... there's really only:
- 2 x temp probes
- 2x float switches
- 2 x wave makers (mp10 x gyre)
- return pump
- skimmer.
- heaters and chiller (both off 90% of time so assume they aren't the culprit)
Sent from my iPhone using Tapatalk
I took the floats switches out.. no change. I will try and isolate cause ... there's really only:
- 2 x temp probes
- 2x float switches
- 2 x wave makers (mp10 x gyre)
- return pump
- skimmer.
- heaters and chiller (both off 90% of time so assume they aren't the culprit)
Sent from my iPhone using Tapatalk
Re: PH probes - reading strange values
Yeah, that's a pretty quick indicator.