PH Calibration
Posted: Sun Dec 14, 2014 5:56 pm
I'm having a little trouble figuring out my PH probe calibration.
Its brand new and I purchased hanna ph 7 and ph 10 buffer solution to calibrate.
I'm getting for pH 7 ~515
and pH 10 ~789
RA is telling me my pH is around 7.12 where my API test kit says 8.3
I tested by putting the probe directly into each of the packets, not sure if that is a problem or not. I let it sit for 10 minutes or more till it stops moving around. Washed with RODI water each time, and never let the probe get dry.
Now everytime i retest i'm getting different numbers on both 7 and 10. The last calibration i put the probe into rodi water and it was saying 10ph on the RA. What am I doing wrong?
Its brand new and I purchased hanna ph 7 and ph 10 buffer solution to calibrate.
I'm getting for pH 7 ~515
and pH 10 ~789
RA is telling me my pH is around 7.12 where my API test kit says 8.3
I tested by putting the probe directly into each of the packets, not sure if that is a problem or not. I let it sit for 10 minutes or more till it stops moving around. Washed with RODI water each time, and never let the probe get dry.
Now everytime i retest i'm getting different numbers on both 7 and 10. The last calibration i put the probe into rodi water and it was saying 10ph on the RA. What am I doing wrong?