Ph drops when jeabo wave maker turned on
Posted: Fri Dec 26, 2014 11:22 am
I am looking for assistance with my pH probe. I recently calibrated the pH probe to 7 and 10, then after calibration and reboot of the controller confirmed that the pH read 7 and 10 in both standards.
However I noticed that the calibration seemed to have drifted. I checked for current leak by turning off individual components. Although I was originally suspecting the skimmer, turns out it is the jeabo wm.
It is causing the ph to go from 8.3-8.4 and drop to 6.7
Question is what should I do to resolve this issue? Is it a grounding issue with the wave maker? Or should I calculate for the electrical interference and manually play with the pH calibration to correct for the interference.
Thanks for any assistance