Posted: Mon Nov 23, 2015 3:52 pm
by Dadof2boys
Here's my situation... I have calibrated my salinity probe with 35 ppt refractive index standard solution and when in the solution the probe is reading 35.0. When I move the probe to my RODI water it reads 6.0 and I know that is really at 0. When I move the probe to my sump the probe then reads anywhere from 34-35.6. I know that my water is really at a 1.025-1.026.

Without just saying take 10 points off the probes value is there a way to force the adjustment?

Re: Manually Calibrate Salinity

Posted: Mon Nov 23, 2015 4:59 pm
by lnevo
6 is the lowest the probe reads. It's normal. It sounds like you're probe is reading perfectly for 1.025-1.026.

Re: Manually Calibrate Salinity

Posted: Mon Nov 23, 2015 6:37 pm
by Dadof2boys
So how do you know what your salinity really is then? Do you just mentally subtract 10 points?

Re: Manually Calibrate Salinity

Posted: Mon Nov 23, 2015 8:00 pm
by rimai
Do you keep your tank at 25ppt?
Salinity measurements are in ppt and not specific gravity.

Re: Manually Calibrate Salinity

Posted: Mon Nov 23, 2015 8:22 pm
by lnevo
35ppt is 1.026