problem with pH reading
Posted: Thu Sep 13, 2012 7:34 pm
After about a month of working correctly, yesterday the pH reading dropped to ~2 and then raised to ~5. I recalibrated the probe. The reading for pH7 and pH10 went down respectively from 554 to 457 and from 832 to 741. When I put the probe back to the tank or sump it still reads ~5. When I test this probe in a cup with the water from the tank it shows 8.11 which is about right. So i thought voltage leaking in the tank. I tried to unplug every item one by one with no results, the reading remained in the ~5 range. Tested it in cup again and it shows over 8 which is correct. What could be the reason of the false reading when the prob is in the tank/sump? Also keep in mind that it just happened yesterday.