ORP probe faulty?

Expansion modules and attachments
Post Reply
butcherman
Posts: 44
Joined: Thu Apr 05, 2012 11:14 am

ORP probe faulty?

Post by butcherman »

My ORP reads find on any reading below 325. at 325 it changes to 0 when ORP drops to below 325 it starts shoing reading again.

Im using 1.0.2 libraries and modified the libraries to calibrate with 220mv fluid.

when i recalibrate the terminator that came with the probe registers 1695
and the 220mv fluid reads 595

Is this the probe, hardaware or the code?
rimai
Posts: 12857
Joined: Fri Mar 18, 2011 6:47 pm

Re: ORP probe faulty?

Post by rimai »

Has it worked in the past?
Roberto.
butcherman
Posts: 44
Joined: Thu Apr 05, 2012 11:14 am

Re: ORP probe faulty?

Post by butcherman »

Yes worked fine previously.

Sent from my GT-I9500 using Tapatalk 4
rimai
Posts: 12857
Joined: Fri Mar 18, 2011 6:47 pm

Re: ORP probe faulty?

Post by rimai »

1695 with the terminator sounds ok, but 595 for 220mV is too low.
Roberto.
butcherman
Posts: 44
Joined: Thu Apr 05, 2012 11:14 am

Re: ORP probe faulty?

Post by butcherman »

Probably best to replace

Sent from my GT-I9500 using Tapatalk 4