PH Calibration and Temperature

Post by Bryan »

Gentlemen:

When calibrating the probe is it temperature compensating. Don't see anything in the code. So should the calibration solutions be kept at close to the tank temp when calibrating, or calibrate with the recommended temp on the solution. ie my 7.0 solution is 7.0 at 77F. or does it really matter?
-=Bryan=-
rimai
Posts: 12857
Joined: Fri Mar 18, 2011 6:47 pm

Re: PH Calibration and Temperature

Post by rimai »

Not temperature compensated
Roberto.
Bryan
Posts: 147
Joined: Wed Jul 06, 2011 1:49 am
Location: Vancouver, B.C.

Re: PH Calibration and Temperature

Post by Bryan »

Thanks, so best to calibrate with the solutions at their recommended temperature.
-=Bryan=-
rimai
Posts: 12857
Joined: Fri Mar 18, 2011 6:47 pm

Re: PH Calibration and Temperature

Post by rimai »

Actually, it's best to calibrate at tank water temperature. Because there is no temperature compensation, you eliminate the chance of variance of pH if there is a difference between standard and solution to be measured.
Roberto.